13. Troubleshooting#

This section contains common troubleshooting tips.

13.1. The Joulescope UI is not working as expected#

If you are using Joulescope UI 1.0 or newer and the UI is still running, select Help → Report Issue. Fill in your name and email so that we can reply to you, then enter a description. Click Submit.

In many cases, you can resolve the issue by clearing the Joulescope UI configuration to revert to default settings. Select File → Clear config and exit, and then restart the Joulescope UI.

If the Joulescope UI 1.0 or newer does not start correctly, you should see an error window:

Joulescope UI error window

Fill in your name and email so that we can reply to you, then enter a description. Click Submit.

Select either Revert to previous configuration or Revert to defaults for the recovery option on the next page.

13.2. JS110 is not recognized by the host computer#

Ensure that the instrument is connected by USB to the host computer. Ensure that the USB connection is directly to a USB port on the host computer. Avoid any USB hubs, docks, or adapters. For desktop PCs, use a motherboard port on the back as some front USB ports use a USB hub internally.

For Linux machines, ensure that you have appropriate permissions. You will usually need to add the udev rules.

13.3. Offset error#

The JS110 is designed to never need recalibration, but it can become damaged. The most common indicator of damage is an offset error in one or more current ranges.

To diagnose this issue, disconnect all cables from the sensor-side IN and OUT. Using the Joulescope UI, manually select each current range and note the offset. Using the Joulescope UI 1.0, select View → Multimeter, then select Widgets → Device Control. Rearrange the widgets. Manually select each Current range in turn, and note the measured current offset at the top left of the Multimeter widget.

Multimeter widget showing the current offset

The current ranges should be within the offset accuracy specifications shown in the current measurement specifications in Section 4. If you find that a current range or two is out of specification, the most likely cause is damage to one of the shunt resistor selection MOSFETs. Known ways to damage these MOSFETs include overvoltage, excessive electrostatic discharge, overcurrent, and reverse current while the JS110 is unpowered.

You can attempt repair of the MOSFET corresponding to the damaged range:

Current Range

Reference Designator

Part Number

18 µA

Q1

ZXMP10A13FTA

180 µA

Q2

ZXMP10A13FTA

1.8 mA

Q3

IRLML2244TRPbF

18 mA

Q4

IRLML2244TRPbF

180 mA

Q5

SI7615ADN-T1-GE3

2 A

Q6

SI7615ADN-T1-GE3

10 A

Q6

SI7615ADN-T1-GE3

After repair, ensure that you remove all solder flux, thoroughly dry the printed circuit board, and allow the printed circuit board to return to room temperature.

13.4. Unable to measure the full voltage range#

You apply a known voltage across IN+ and IN-, but the JS110 measures a smaller voltage. First ensure that you have the Voltage range set to 15 V. If you are using the Joulescope UI 1.0, you can find this setting in the Device Control widget.

If the JS110 is set to the 15 V current range, then the JS110’s Ćuk converter may not be generating a sufficient voltage. The most likely cause is that one of the capacitors failed. Open the JS110 and measure the voltage across C116 or C120, which are the two capacitors in the Ćuk converter that seem to fail more than they should. You should see -16V.

When the JS110 is powered off, you can use a multimeter to measure their resistance across these capacitors. You should measure 10 kΩ or more in circuit. Note that you may need to wait for the multimeter to charge up the capacitors to make this measurement.

If the voltage is not close to -16 V, or the resistance measured too low, then you can attempt repair by replacing them. They are 4.7 µF 1206 (imperial) 50V 20% X5R. One suitable part is UMK316BJ475KL-T.

13.5. Unable to open in Joulescope UI#

In Windows, the device appears in Device Manager. However, when opening the Joulescope UI, the device does not connect. In the Device Control widget, the “Open and close device” circle is not available and grayed out.

The most likely cause is a fault on the sensor-side of the JS110. When the UI attempts to open the device, it powers on the sensor-side. If the sensor-side consumes too much power due to a hardware failure, then the host software will time out while trying to initialize the JS110’s sensor-side.

The normal failure is that U33 becomes hot. U33 is a TI TPS2552 current limiter which drops the voltage and dissipates power on overcurrent conditions.

The most common failure that causes overcurrent are the Ćuk converter capacitors C116 and C120. See Section 13.4 for troubleshooting and repair.

13.6. Inaccurate measurements in the presence of strong RF#

The JS110 is susceptible to strong RF. The JS110 is a shunt ammeter that maintains a maximum of 20 mV across the shunt resistor. Nearby transmitters can easily induce much larger voltages. For example, 30 dBm = 1 W = 20 V peak-to-peak in a 50 Ohm transmission line. Even 20 dBm = 0.1 W = 6.324 V peak-to-peak. While the JS110 does have some input filtering, the filtering is very limited and the front-end opamp has no internal EMI attenuation.

The JS110 is also an unbalanced instrument, meaning that its measurement range is biased towards positive measurements. For the same current range, the positive measurement limit is about five times the negative limit. High frequency signals will saturate the JS110’s range, which clips at the measurement extremes rectifying the RF signal. When this situation occurs, the JS110 will always measure higher current than actual.

The JS110 is often used with target devices that are RF transmitters. To minimize RF pickup, we recommend the following:

  • Twist together the wires going to IN+ and OUT+ to minimize the loop area. This minimizes coupling from the antenna to the JS110 current sense wires.

  • Keep the JS110 at least 1 wavelength away from the antenna to minimize near-field coupling directly to the JS110 enclosure.

  • Keep the host-side USB connection away from the antenna, too.

  • For more accurate current measurement, connect IN+ and OUT+ as a twisted pair directly from your target device. Consider placing a ferrite bead at each end of the twisted wire.

  • Consider upgrading to the Joulescope JS220 which is a balanced instrument (equal + and - limits) and has further filtering improvements to minimize RF interference.