14. Maintenance#

The Joulescope JS320 is designed for minimal maintenance. To clean your Joulescope, disconnect all wires from the unit. You can gently wipe the unit with a damp cloth using water only. Using rubbing alcohol or other cleaning agents may damage your Joulescope. Do not submerge.

Joulescope is factory-calibrated and should continue to function within specifications over its operating life. For critical or demanding applications, you may choose to follow industry-standard practices and have your Joulescope recalibrated annually. Tra-Cal, located in Maryland, USA, provides authorized ISO 17025:2017 calibration and recalibration. You can submit your request at:

https://www.tra-cal.com/request-a-quote

Recalibrating Joulescope requires custom equipment and trained personnel. Only Jetperch LLC and Tra‑Cal are authorized to perform recalibration.

14.1. Verification Procedure#

Calibration should normally find that the Joulescope instrument still operates within specification. Although calibration can only be performed by authorized vendors, any facility capable of performing precision multimeter calibration can verify that a Joulescope is operating within specification. This section describes the verification procedure.

14.1.1. Preliminary operations#

  1. Download the host computer software from https://www.joulescope.com/pages/downloads

  2. Install the software

  3. Download the validation spreadsheet from https://download.joulescope.com/docs/validation/index.html

  4. Connect your Joulescope to the host computer using the provided USB cable.

  5. Select View. If the first Multimeter is not already selected, select it.

  6. Select View. If Control is not already selected, select it.

  7. Open the validation worksheet in Excel and fill in the Device Under Test Information.

  8. Select File → Save As for this device calibration.

  9. Optionally configure your Test Uncertainty Ratio (TUR) to your desired level. We recommend following industry-standard practices and set TUR to 4 or greater.

14.1.2. Voltage Validation Process#

  1. Connect calibrator OUT terminals to the Joulescope Voltage terminals, observing polarity.

  2. Connect Voltage + to Current +. Leave Current – open.

  3. In the Joulescope UI, select Voltage Range to 15V.

  4. Configure the calibrator to the target voltage.

  5. Record the actual calibrator voltage, the calibrator’s uncertainty at that voltage, and the Joulescope’s measured voltage as displayed by the Joulescope UI.

  6. Repeat 4 and 5 for each target voltage.

  7. Repeat 2 through 6 for the 2V range.

The following image shows the calibration spreadsheet:

Voltage validation spreadsheet for the 15 V and 2 V ranges

During the test process, you should fill in all light-blue cells with measurement data. All TUR values should be green. Yellow indicates that the test setup lacks sufficient accuracy. All “As Measured” values should be green. Red indicates that the Joulescope does not meet specifications.

14.1.3. Current Validation Process#

The current calibration process can be performed using several methods. In all cases, the goal is to generate a current close to the target current. Strict accuracy on the generated current is not required. However, a reference measurement is required.

The general process is

  1. Select the Current Range in the Joulescope UI.

  2. Configure the target current.

  3. Record the applied current as measured by the reference, reference’s uncertainty at that current, and the Joulescope’s measured current as displayed by the Joulescope UI.

  4. Repeat 2 and 3 for each target current

  5. Repeat 1 through 4 for each current range.

Current validation spreadsheet for each current range

Note that many of the test cases are for a 0.0 amp target current. In this setup, you can simplify the setup and leave the Joulescope instrument disconnected, or optionally short the Joulescope Current terminals with a banana jumper wire.

Numerous valid test equipment methods exist to generate and accurately measure the target current. You may select a method that is appropriate for your equipment. The Joulescope JS320 has a very large current range, and you will need to accurately measure from 18 µA to 3 A. You may need to employ more than one method to cover the entire current range with a sufficient TUR. Possible methods include:

Method 1: Reference current source

If you have a reference current source, connect the current source to the Joulescope Current terminals, observing polarity.

In step (2), directly configure the reference current source to generate the target current.

In step (3), read the applied current from the reference current source.

Method 2: Reference ammeter with current source

Connect an inaccurate current source + terminal to the Joulescope Current + terminal. Connect the Current – terminal to the reference ammeter + terminal. Connect the reference ammeter – terminal back to the current source – terminal.

In step (2), configure the inaccurate current source to the target current.

In step (3), read the applied current from the reference ammeter.

Method 3: Reference ammeter with voltage source and variable load resistor

Connect an inaccurate voltage source + terminal to the Joulescope Current + terminal. Connect the Joulescope Current – terminal to one terminal of a variable load resistor. Connect the other variable load resistor terminal to the reference ammeter + terminal. Connect the reference ammeter – terminal to the Joulescope inaccurate voltage source – terminal.

In step (2), select a target voltage and resistance combination to give the target current as computed using Ohm’s law: I = V / R. Configure the load resistance, configure the voltage source, then enable the voltage output. Ensure that the voltage stays between 0 and 15V. We recommend adding a step 3.5 to disable the voltage output before changing the load resistance for the next target current.

In step (3), read the applied current from the reference ammeter.

14.2. Offset Calibration#

The JS320 is designed to remain in spec without calibration. However, you can perform offset calibration to improve accuracy for demanding measurements.

The JS320 supports offset calibration for both current and voltage. The Joulescope UI 1.0.40 and later includes offset trim calibration. Before starting a trim calibration, please run your JS320 for at least 5 minutes to allow the temperature to stabilize. To start a current offset trim calibration, select the Device Control widget in the sidebar, then press Calibrate:

Joulescope UI calibration dialog

Press Start for Current offset or Voltage offset, and follow the onscreen instructions. Ensure that you leave the terminals open for Current offset calibration. Short V-, V+ and I+ to perform Voltage offset calibration.

The JS320 uses an internal fixed-point representation of 5Q31 for current samples. 2⁻³¹ represents an accuracy of ±0.5 nA. After performing trim calibration, the instrument offset may be off by up ±0.25 nA.